Ian Morgan
Online HDF fluid
Posted on: 08/11/2007 16:24:54

Dear All,


I am looking for some advice on how to test HDF reinfusion fluid. I only have experience of Gambro ‘Ultra’ machines & Nikkiso DBB-05’s. Both of these filter the dialysis fluid internally before passing the ultrapure fluid through a disposable single use endotoxing filter. We routinely test the output of the machines (without the single use filter) to ensure it complies with the current RA guidelines; TVC <0.1cfu/ml and endotoxin <0.03 IU, we send the lab. 500ml of the fluid so they can filter and culture for 7 days on TGEA; we occasionally have a TVC of 4 or 6 in 500ml (TVC 0.012 cfu/ml) but mostly it is zero. We have the Fresenius 5008 here at King’s to trial and I am not sure how to sample as this machine does not use a single use filter, only the internal one; therefore the RA guidelines for this fluid say it must be sterile (<1 cfu/1000 litres). Is a 500ml sample enough, and clearly even 1 cfu in 500ml would be unacceptable. Having read the 5008 operator manual it states that the unfiltered dialysate must have a TVC of <1000 cfu/ml and the ‘High-purity dialysate’ must have a TVC of <1 cfu/ml. If you follow this through this recommendation it would mean it is acceptable to have a cfu of 9/10ml in the reinfusion fluid. When you multiply this up to a substitution volume of 20 litres it would mean you have infused into the patient enough bacteria to produce 18,000 colony forming units! Any advice gratefully received.


Ian

 
Fraser Gilmour
Re: Online HDF fluid
Posted on: 15/11/2007 09:33:24

Hi Ian,


Having introduced HDF therapy with the 5008 at one of our units in Dorset earlier this year, we have been through the same problems with validation of the infusion fluid. The RA guidelines are quite clear about the limits as you say, less than 1 in 1000L and 0.03EU/ml. The problem this gives is that the only way to be 100% sure that the fluid meets this standard is to filter 1000L and see no growth, the direct inoculation method using 1ml would not give a high enough confidence level.


This is obviously impractical, so after talking to the microbiologist at our lab we agreed that they would filter a 1L sample (1L being the most we could practically draw off regularly and that they could process) taken from the 5008 safeline directly into a sterile prime bag. This was done with random machines regularly over a three month period to cover ultrafilter changes, we also sent control samples of fluid that would not meet this standard.


All infusion fluid samples came back showing no growth in 1L and less than 0.03 EU/ml (I would have liked do quantative endotoxin down to 0.01 but our lab could not do it at the time).


This validation period coupled with the 5008 ultrafilter integrity test has given us the confidence to stop regular testing of the infusion fluid, we now only take 100ml samples of water from septum ports at the machine inlet to check they meet ultrapure standard. Chemical analysis is still done three monthly.


As there are no further guidelines from the RA, this is what we felt comfortable with. Hopefully the next revision of the guidelines will be more explicit.


BR


Fraser

 
Ian Morgan
Re: Re: Online HDF fluid
Posted on: 15/11/2007 16:25:59

Hello Fraser,


Thanks for replying to this posting. Since then I have done some literature searches and come up with more details as well as speaking to a few other users. There appears to be no consensus about how much fluid should be collected when testing the substitution fluid with, some units collecting 10ml, and others collecting 100ml; whilst Fresenius in the UK only use 1ml when testing this fluid in their laboratory.


The EDTNA/ERCA Guidelines; Quality assurance for dialysis-quality water and dialysis fluid site a French document, Circulaire DGS/DG/AFSSAPS No 311 which has now been replaced with the following standard: http://www.sante.gouv.fr/adm/dagpb/bo/2007/07-03/a0030058.htm


This recommends collecting 100ml of ultrapure dialysis fluid and ensuring the TVC is <10cfu (<0.1cfu/ml). For the substitution fluid it recommends collecting a minimum of 500ml and ensuring there are 0cfu. This standard also recommends making these measurements once every three months.


This makes sense as a smaller volume cannot guarantee to show very low bacterial levels. Whilst this does not meet the stringent requirement needed to demonstrate sterility as defined by the European Pharmacopoeia (this requires 10% of a container be filtered and cultured for 14 days) it is achievable. This has reassured me that they processes we currently use are adequate for testing on-line HDF machines without a single-use filter.


Best regards


 


Ian

 
Gareth Murcutt
Re: Online HDF fluid
Posted on: 16/11/2007 08:38:01

Hi
ART were given the chance to comment on the latest draft of the RA guidelines after some of us pointed out to them that one cfu per tanker load was quite an exacting stabndard. Lizzie helped to pull together an excellent revision which I believe was accepted. I thought it had been published here or was available on RA site. I'll check and get back to you.
Gareth

 
Gordon Farquhar
Re: Re: Online HDF fluid
Posted on: 16/11/2007 08:51:44

Hi


Lizzie's full article is available in our Document Index under Artery Newsletter Issue 41.


Gordon

 
Gareth Murcutt
Re: Re: Re: Online HDF fluid
Posted on: 16/11/2007 08:59:23

Latest version of RA Clinical Practice Guidelines are available at http://www.renal.org/guidelines/module3a.html#Concentrates_Water. There is also a good rationale section below the actual guidlines. I haven't had a chance to read it through but noted that it said HDF fluid guidelines up for review in next edition (2009). Volunteers to help them draft it?

 
Ian Morgan
Re: Re: Re: Re: Online HDF fluid
Posted on: 16/11/2007 12:38:47

Maurice has just asked if this document could be posted http://jasn.asnjournals.org/cgi/content/full/13/suppl_1/S78


In reply:


Hello Maurice,


Thanks for this; though I have already seen it and I have to say it did not help much when trying to validate a practical program of monitoring. Obviously this paper is biased to showing how safe the Gambro system is: my concern was how to test a system without a single use filter. Here they validated the sterility of the final infusion fluid by collecting 2 x 15 litres which is not practical for us. I think the information that needs to be got across is that you must collect as large a volume as is practical, and I don’t think 10 or 100ml is acceptable. I am currently waiting for information from Fresenius to see how they validated their process, but I bet they too used larger volumes than we could hope to send to our labs on a routine basis.


Best regards


Ian

 
Gareth Jones
Re: Re: Re: Re: Re: Online HDF fluid
Posted on: 21/12/2007 17:07:00

I have asked Rolf Nystrand to offer some opinion on the matter and he has provided me with a document that I have today posted on the Gambro UK website, in the Product Section, In-Centre section.  There is also a link on the Home page, product news.


I don't pretend to even begin to understand what he's written and it's way beoynd my expertise, but I hoped to be able to contribute with a supplier's perspective.


If you have further, questions, please don't hesitiate to contact Rolf directly, or you can send me an e-mail and I'll try to get you a repsonse.


Merry Christmas


Gareth Jones


gareth.jones@gambro.com